Roberto,
This is likely a case where a custom designed solution would be best, but this also has to be a solution that is practical given the likely budget and potential volume of the application.You will need to have a copper or aluminum plate near the dimensions of the heater in order to achieve thermal uniformity.
The HK5572 R4.1 L12B (0.5 x 0.5in) and HK5578
R5.4 L12B are the smallest square catalog heaters in the 1-2cm2 range that Minco has available. These also have hi-TCR foil options for use with a CT198, which might be the most practical way to maintain the temperature.
You will probably want a thermistor based temperature indicator which could detect 0.1C increments over the +10C to +50C range. The thermistor would need to be able to be mounted in the metal plate to ensure it is actually sensing the temperature you want rather than the heating element. You would need to adjust the CT198 setpoint until the thermistor indicator unit shows the temperature of interest has been reached.
Minco would be limited to the supply of PSA-backed (catalog) heaters and the CT198 controller. You would need to provide copper plate; sensor; voltage source; voltmeter,
ohmmeter, temp readout, etc.; and perform the temperature measurements to calibrate the system.